April 20, 2024, 08:58:58 AM
Forum Rules: Read This Before Posting


Topic: Analyzing aluminum metal for strontium  (Read 6733 times)

0 Members and 1 Guest are viewing this topic.

dwebb210

  • Guest
Analyzing aluminum metal for strontium
« on: July 15, 2005, 04:29:25 PM »
I guess I have to admit defeat.

I work at a metallurgical lab, and the latest puzzler is
a bunch of aluminum samples that need to be analyzed for
strontium content.

Our optical emission spectrophotometer can't detect
strontium, so I have to digest the sample and run it
on a flame AAS.

The aluminum has around 8% silicon.

I don't have metal chip standards with known strontium values.

I do, however, have a 100ppm multielement standard, which
includes strontium.

I know aluminum and silicon both interfere with strontium
analysis, so matching the matrix is important.

I took six flasks and weighed out one gram of chip standard
into each.  I added 0.1, 0.2, 0.5, 1.0, and 5.0 milliliters
of the multielement standard into five of the flasks, and
left one as a "blank".

After digestion, the solutions were diluted to 100mls,
and filtered.

I was essentially doing a "standard addition" test, because
not only did I end up with a set of standards, I was able to
calculate how much strontium was in the metal chip standard.

Or so I thought.

I digested the 6 standards in nitric acid and hydrofluoric.
The reason was because I wanted to save time and avoid
the need to filter the samples if possible.
The use of HF got the silicon and titanium into solution...
but a white precipitate formed which needed to be filtered anyway.

I digested one of the unknown samples the same way.

Realizing I wasn't going to save any time by using HF,
I tried digesting another 6 standards with the Sr spikes,
only this time instead of HF, I used nitric and HCl.
I filtered out the undissolved Si and Ti.

I digested the same unknown sample this way too.

So I have a set of standards with HF, and a sample with HF.
And a set of standards with HCl, and the same sample with HCl.

I ran everything on the AA.

I corrected for the weight of the metal chips,
I corrected for what appeared to be strontium in the chips,
I'm sure I did the calculations right.

I got linear calibration curves for both sets of standards,
although the instrument response was roughly three times
greater for those digested with HF.  
The correlation coefficient for both was >0.999

I expected a difference in instrument response,
and I was happy that I got a linear calibration curve for
both types of dissolution.

HOWEVER, the sample dissolved with HCl calculates to 0.011% Sr,
while the sample dissolved with HF calculates to 0.028% Sr.

Strontium is extremely important for this alloy, and a difference
as large of this is unacceptable.  

Where did I go wrong?

The only thing I haven't done yet is digest another two
samples to verify the values.  There is a remote possibility
the drill hit an inclusion that contained high strontium,
and the drilling made it into one flask but not the other.

Does anyone have any suggestions?

Thanks

Dave


Offline Borek

  • Mr. pH
  • Administrator
  • Deity Member
  • *
  • Posts: 27652
  • Mole Snacks: +1800/-410
  • Gender: Male
  • I am known to be occasionally wrong.
    • Chembuddy
Re:Analyzing aluminum metal for strontium
« Reply #1 on: July 15, 2005, 05:43:26 PM »
Please post in appropriate forums, this is analytical chemistry question.

Perhaps you are loosing part of strontium with precipitate? Which samples had more precipitate/unsolved material filtered out? These digested in HNO3/HCl? If so, that could explain weaker signals in these samples.

Just a wild guess.
« Last Edit: July 15, 2005, 05:45:25 PM by Borek »
ChemBuddy chemical calculators - stoichiometry, pH, concentration, buffer preparation, titrations.info

Sponsored Links