I apologize for the long post. I’m of the opinion that less is more, but I was at a loss to include all of the pertinent information in a more succinct manner. It’s the best I could do. I sincerely thank any and all for their help, thoughts, criticisms, and/or advice.
I’m performing a GC-MS analysis on a complex sample. I can safely assume no matrix interferences, but it is a complex sample with many products of interest that are homologous in nature—such as a series of fatty-acid derivatives. My standard contains over a dozen of these derivatives in methylene chloride and is dilute. I can run the standard as is with a 20:1 split with peak abundances no lower than 200-300k and no higher than 1 billion. In other words, it’s not concentrated, so I wouldn’t, typically, consider using it as a “stock solution.” I only have one mL of this standard that comes in an ampule and it’s expensive.
Brass tacks: I’m limited in my standard by both amount and cost, and it’s tricky to work with once opened, since the solvent is volatile. I want to run a 5-point calibration curve, but I don’t think I can work quickly enough to generate a good curve. At absolute best, I could perform a splitless analysis and dilute my standard 1:20 and use that to prepare 5 standards. But the appropriate solvents are all so volatile (methylene chloride, pentane, carbon tet, etc.) that even a 1:20 dilution will be hard to prepare accurately, let alone use of that solution to prepare additional standards.
Hence, the heart of my question follows. What if I prepared a dilution of my standard fresh from the ampule. Say, a 1-to-5 dilution. Then, I inject 1 mic. Run it again but inject 2 mics. Run again but inject 3 mics. And then 4 mics and 5 mics. If we call the 1:5 dilution “soln. A,” my scheme would look like this:
Standard 1: soln. A – 1 mic injection
Standard 2: soln. A – 2 mic injection
Standard 3: soln. A – 3 mic injection
Standard 4: soln. A – 4 mic injection
Standard 5: soln. A – 5 mic injection
In theory, the MS is a mass-dependent detector and not concentration-dependent. Injecting 2 mics of a 5 ppm standard would, in theory, inject the same mass of product as 1 mic of a 10 ppm standard. In theory. In practice, is this okay? I’m aware that there could be confounding issues here. This is REALLY just a test of the linearity of my injector, but, IN THEORY, the calibration curve would be the same as a traditional calibration curve, assuming reproducible injection behavior and no confounding other issues. I also realize that, yes, an MS is mass-dependent, but there could be other concentration-dependent variables that affect the signal. It’s simply the difference between theory and practice.
Finally, here’s my ultimate question: for those of you with more extensive GC-MS experience, can I get away with this? I realize it depends a LOT on the nature of my sample and standard, but is this “trick” commonly employed in practice or is it, rather, seen as inadvisable?
I don’t need spectacular error bars, but I do need better than “ballpark” results. Lastly, as I’m sure some of you are considering while reading this: reproducibility of injection. The Achilles heel to my plan (or one of them) is that I’d be unable to employ an internal standard to correct for injection variability. After all, if I am constantly changing injection volume, I can’t rely on the signal from an internal standard to remain constant and normalization goes out the window. However, given other complications, no matter how I perform this analysis, I will not be able to employ an internal standard, anyway. I will, however, be able to inject each “standard” (from the outline above) and each sample more than once. I intend to run every solution in triplicate. I also know from extensive use with this GC-MS that its ALS is about as good as it gets. It’s not perfect, but it’s as good as it gets.
I’m not shy, and I’m not insecure. If this is a bad idea, please let me know. I’m open to all criticisms. Could this potentially work, or am I simply neglecting too many potential variables that would alter the nature of my calibration curve?