April 20, 2024, 03:16:44 AM
Forum Rules: Read This Before Posting


Topic: ESI and tandem in space mass spec?  (Read 2210 times)

0 Members and 1 Guest are viewing this topic.

Offline dmf10006

  • Very New Member
  • *
  • Posts: 1
  • Mole Snacks: +0/-0
ESI and tandem in space mass spec?
« on: September 17, 2014, 03:48:15 PM »
Hi Everyone. So, I am having a bit of trouble figuring out how electron spray ionization and MS/MS work together to give data. My first problem is coming from most likely my lack of understanding of the entire process from injection to the final data obtained after ion fragments are analyzed in MS2. Another issue I'm discovering is how would I know whether I have stable molecular ions from electron spray or fragmented ions. Just due to my limited understanding, layman's terms would be greatly appreciated.

Offline MOTOBALL

  • Full Member
  • ****
  • Posts: 371
  • Mole Snacks: +50/-5
Re: ESI and tandem in space mass spec?
« Reply #1 on: September 18, 2014, 02:39:35 PM »
The advantage of ESI (electrospray ionization) is that it is a low-energy process (so-called "soft" ionization) that ionizes the analyte of interest with a minimum of excess energy deposited in the neutral molecule.  The mass spectrum will therefore, IN GENERAL, show only [M+H]+ together with ions formed by loss of small neutrals (NH3, H2O, MeOH, EtOH etc) from [M+H]+.

The disadavantage....etc. etc.

If you need structural information, the [M+H]+ ion needs to be further fragmented (in addition to losses described above), by putting more energy into the protonated molecule.  This is usually achieved by CID (collisionally induced dissociation) also known as CAD (collisionally activated dissociation), by selecting the [M+H]+ ion in the first mass analyzer (first stage MS) and passing ions of this m/z value ONLY into a collision cell that is pressurized with an inert gas (N2; Ar, usually) at low pressure.  Collisions result in the fragmentation of [M+H]+ to produce structurally significant ions,  which are transmitted from the collision cell to a final mass analyzer for separation (second stage of MS).

MS/MS means two stages of mass analysis (abbreviated as MS2 in ion-trap instrumentation), as described above.

 how would I know whether I have stable molecular ions from electron spray or fragmented ions

The [M+H]+ ion will often (BUT not always) be accompanied by [M+Na]+ and [M+K]+ signals, usually of MUCH lower intensity than the [M+H]+ signal.


Sponsored Links