Trouble with mobile phase in HPLC
« on: February 01, 2016, 12:25:04 PM »
Trying to investigate lidocaine in wastewater but haven't found a peak on chromatogram when running through a stock lidocaine standard. Have used methanol/water mixtures and phosphate buffer(pH 4)/acetonitrile ones.  Any ideas on how to proceed?

Offline Babcock_Hall

  • Chemist
  • Sr. Member
  • *
  • Posts: 5609
  • Mole Snacks: +321/-22
Re: Trouble with mobile phase in HPLC
« Reply #1 on: February 01, 2016, 04:39:14 PM »
What wavelength of detection are you using, and why did you choose it?  What do you know about the solubility of lidocaine in the mobile phases you have chosen?

Offline snov

  • Very New Member
  • *
  • Posts: 2
  • Mole Snacks: +0/-0
Re: Trouble with mobile phase in HPLC
« Reply #2 on: February 01, 2016, 05:23:30 PM »
210 and 260nm on seperate runs. Just going by past examples seen in journals.  I have read that it is mostly insoluble in water and soluble in alcohols. Log Kow is ~ 2.4.

Offline Arkcon

  • Retired Staff
  • Sr. Member
  • *
  • Posts: 7367
  • Mole Snacks: +533/-147
Re: Trouble with mobile phase in HPLC
« Reply #3 on: February 01, 2016, 06:40:59 PM »
Many more things are possible other than mobile phase.  Are you exactly coping journal procedures and seeing nothing?  Is your system working properly for other analyses?
Hey, I'm not judging.  I just like to shoot straight.  I'm a man of science.

Offline Babcock_Hall

  • Chemist
  • Sr. Member
  • *
  • Posts: 5609
  • Mole Snacks: +321/-22
Re: Trouble with mobile phase in HPLC
« Reply #4 on: February 01, 2016, 07:00:52 PM »
I agree with Arkcon, but I have one more question.  How are you preparing the standard?

Offline kez8zek

  • Very New Member
  • *
  • Posts: 1
  • Mole Snacks: +0/-0
Re: Trouble with mobile phase in HPLC
« Reply #5 on: February 18, 2016, 09:50:03 PM »
dissolve standard in 60:40 ACN Water. try mobile phase 60% ACN: 40% 25-50mM phosphate buffer ph=6-7. Should be RT<10min

Offline Arkcon

  • Retired Staff
  • Sr. Member
  • *
  • Posts: 7367
  • Mole Snacks: +533/-147
Re: Trouble with mobile phase in HPLC
« Reply #6 on: February 18, 2016, 10:35:56 PM »
This method here is from a column manufacturers website:  https://www.phenomenex.com/Application/Detail/3254?returnURL=/Compound?id=Jetocaine&alias=Jetocaine

That method is different than yours, but not too far away.  If you have that column, you might want to try it.  What level are you at?  How much are you injecting on-column, and can you see something if you inject more?  Are you sure your system is ok? Can you run caffeine, or other test mixture to see that the system is working properly?
Hey, I'm not judging.  I just like to shoot straight.  I'm a man of science.

Offline Babcock_Hall

  • Chemist
  • Sr. Member
  • *
  • Posts: 5609
  • Mole Snacks: +321/-22
Re: Trouble with mobile phase in HPLC
« Reply #7 on: February 19, 2016, 10:07:02 AM »
Lidocaine has a tertiary amine that should protonate in the pH range of your buffer.  My thought would be to use buffer in the preparation of the standard also (so as to protonate the lidocaine), but I will defer to others with greater knowledge of chromatography.