April 19, 2024, 05:31:45 PM
Forum Rules: Read This Before Posting


Topic: Validating LoQ (Reporting Limit) for FTIR Method  (Read 1885 times)

0 Members and 1 Guest are viewing this topic.

Offline JWF239

  • Very New Member
  • *
  • Posts: 2
  • Mole Snacks: +0/-0
Validating LoQ (Reporting Limit) for FTIR Method
« on: January 11, 2018, 11:23:56 AM »
Hello,

Hoping to find someone with extensive experience in FTIR method development to give me an expert opinion on this question:

Can you verify a quantitation limit in FTIR analysis without specifically having a standard at that reporting limit? Or would you only accept a quantitation limit (or reporting limit) if you had specifically analyzed a standard at that concentration?

I don’t want to get too far into it, but I used two methods from ISO 12828-1 “Validation Method for Fire Gas Analysis”. Main method 1 – Determination of LoD and LoQ from matrix data from blank samples, and Main method 2 – Determination of LoD and LoQ from the linearity of calibration data.

There is also a main method 3 – checking a given or prescribed quantification limit, but it is difficult to obtain gas standards at the concentrations that I calculated from main method 1 and 2.

My understanding was that since FTIR measures the entire spectral range during each scan, even “blank” analysis would yield a signal and therefore having a standard specifically at the desired reporting limit was not strictly necessary, or is this incorrect? Would you approve a method where the reporting limit was calculated using the main methods 1 and 2 above, but the limits had not been explicitly tested with a standard at that concentration, or would you not use a method without having tested a reporting limit specifically? Note the reporting limit is well below the specification limits for my desired commodity specifications. Those limits have been verified with a standard at those limits.

Sponsored Links