In trying to identify well resolved unknown peaks on a liquid sample GC (mostly hydrocarbon + O + Cl) what are the relative merits of GC-MS versus NMR? I do have some idea of the expected class of molecular targets.

Does NMR always need pure cuts to be prepared? I mean, in the case of a GCMS the GC does this separation job; what's the analogous separative stage for a NMR?

In terms of ease of interpretation or uniqueness of a hit is there a tradeoff? How does one select GCMS vs NMR for general sample identification work.

Title: Re: GC-MS versus NMR
Post by: Archer on October 21, 2013, 09:29:21 AM
NMR does not have to be pure, it is very helpful if it is but provided that the impurities do not interferre with the sample peaks it is not a problem.

Usually it is prudent to run GC-MS first as this provides an idea of purity and some idea of what the impurities are.

I don't have an NMR but I do have access to one, I can usually get by with GCMS alone provided that reference materials are available.

Aliphatic hydrocarbons are notoriously tiresome by 1H NMR, particualrly if they are long chains (even with 500MHz instruments). All Upfield signals, all in the same area. If your impurities are all similar to your compound of interest then NMR may not give you the information you need (unless you are looking for a specific diagnostic peak which is upfield from the C-H's).

Analogous stage would be preparative chromatography. The nice thing about GCMS is you only a tiny sample whereas NMR you need (ball park) 1mg / C of sample for a decent 13C spectrum.

There are LC-NMR systems available but these require expensive mobile phases and only provide 1H NMR data.
Title: Re: GC-MS versus NMR
Post by: curiouscat on October 21, 2013, 09:30:07 AM
It might be relevant that a lot of the targets of interest are in the 0.1% to 5% range.

I guess not a major constraint since I could always flash out some lights to concentrate the molecules of interest, within limits.
Title: Re: GC-MS versus NMR
Post by: discodermolide on October 21, 2013, 09:31:44 AM
Usually for NMR one uses pure individual components. Purified by chromatography or crystallisation.
GC-MS gives you the MWt. of each component but no structural info. as is the case for NMR.
Some GC-MS packages come with a large compound database so it may spit a structure out at you.
Ideally one uses them in tandem.
For general identification the GC-MS with database may be an option for you.
Title: Re: GC-MS versus NMR
Post by: curiouscat on October 21, 2013, 09:36:36 AM
@Archer: Thanks for the advice!

Sample quantity, luckily, is not a problem. We've plenty of sample.

Pure cuts are another story. Those are not available.

Reference samples are also hard to get. We only have access to a few.

When you write that for an NMR " impurities do not interfere with the sample peaks" how does one quantify interference. In the sense, if my sample shows 5 unknown peaks on a GC at the 1% range, is it feasible to inject a unseperated sample into a NMR and expect identification of all?
Title: Re: GC-MS versus NMR
Post by: curiouscat on October 21, 2013, 09:38:08 AM
@disco: Thanks!

GC-MS gives you the MWt. of each component but no structural info. as is the case for NMR.

In addition the the MW, doesn't the m/z peak pattern count as structural info?

Or is it too close to call?

Our experience with " compound databases" in the past hasn't been great. They had too many spurious matches.
Title: Re: GC-MS versus NMR
Post by: Archer on October 21, 2013, 09:48:05 AM
When you write that for an NMR " impurities do not interfere with the sample peaks" how does one quantify interference. In the sense, if my sample shows 5 unknown peaks on a GC at the 1% range, is it feasible to inject a unseperated sample into a NMR and expect identification of all?

Non-interferring peaks means that they are resolved from the peaks of interest.

1% should not be too much of an issue even if they are not resolved even 5% in total may not cause you too many issues. Sometimes it is best to run it and find out.


In addition the the MW, doesn't the m/z peak pattern count as structural info?

It will give some structural info (fragmentation patterns etc) but it is of limited use in, for example, determining positional isomers on aramatic rings etc.
Title: Re: GC-MS versus NMR
Post by: DrCMS on October 21, 2013, 12:04:42 PM
When you write that for an NMR " impurities do not interfere with the sample peaks" how does one quantify interference. In the sense, if my sample shows 5 unknown peaks on a GC at the 1% range, is it feasible to inject a unseperated sample into a NMR and expect identification of all?

No.  If you have a sample made up from many different materials all at low levels the NMR will just be a mess of peaks. 

I thing for your application you either go for GC-MS (you may well need accurate mass MS or GC-MS-MS to get sufficient structural info to assign each compound without a library fragmentation match) or you attempt some other separation technique before giving purer cuts a go by NMR. 
Title: Re: GC-MS versus NMR
Post by: curiouscat on October 21, 2013, 01:05:01 PM
Thanks @DrCMS.

Out of curiosity, has a GC-NMR never been an instrumental possibility? Giving the same separative advantages a GC gives an MS?

Title: Re: GC-MS versus NMR
Post by: discodermolide on October 21, 2013, 01:06:33 PM
@disco: Thanks!

GC-MS gives you the MWt. of each component but no structural info. as is the case for NMR.

In addition the the MW, doesn't the m/z peak pattern count as structural info?

Or is it too close to call?

Our experience with " compound databases" in the past hasn't been great. They had too many spurious matches.

The ionisation pattern will point you in a direction leading towards the structure, but it will not give you all you need. NMR is the tool for that, combine the two and you should get what you want to know.
Title: Re: GC-MS versus NMR
Post by: discodermolide on October 21, 2013, 01:08:11 PM
Thanks @DrCMS.

Out of curiosity, has a GC-NMR never been an instrumental possibility? Giving the same separative advantages a GC gives an MS?

GC-NMR good idea, but I think the sample amounts would be too small for the NMR. LC-NMR was also available, but I'm not sure of its status these days.
Title: Re: GC-MS versus NMR
Post by: curiouscat on October 21, 2013, 01:20:57 PM
In terms of functional group identification, what's a good instrumentation strategy.

i.e. Say I've a 2% by product peak on the GC and it may or may not contain a Cl and I want to know if it does. Or similarly to know if it has an OH group or not at all.

Is there any preferred strategy here? In this case, not aiming for full structure elucidation but only pragmatically some smaller goals.

Does FT-IR help? Or is GC-MS / NMR assisted full structure resolution the only way to go.
Title: Re: GC-MS versus NMR
Post by: discodermolide on October 21, 2013, 01:39:15 PM
GC-MS will certainly provide the isotopic content, that's characteristic of each element, like Br or Cl.
Depending upon ionisation mode differentiation between functional groups may be difficult. I still say you need an NMR somewhere in the picture for accurate functional group identification. IR is rather limited for various carbonyls, OH and aromatics for C-Cl or C-F I don't think IR gets you very far.
.
Title: Re: GC-MS versus NMR
Post by: Archer on October 21, 2013, 02:50:09 PM
When you write that for an NMR " impurities do not interfere with the sample peaks" how does one quantify interference. In the sense, if my sample shows 5 unknown peaks on a GC at the 1% range, is it feasible to inject a unseperated sample into a NMR and expect identification of all?

No.  If you have a sample made up from many different materials all at low levels the NMR will just be a mess of peaks. 


I disagree, I have run crude samples, the proton may have a lot of baseline material but the major compound will be 95% of the sample. This may still provide useful data, particularly in HSQC and HMBC experiments where the attenuation can be adjusted to eliminate the small peaks.

It is all very dependent on your substrate and the related impurities though.
Title: Re: GC-MS versus NMR
Post by: Corribus on October 21, 2013, 03:03:42 PM
Out of curiosity, has a GC-NMR never been an instrumental possibility? Giving the same separative advantages a GC gives an MS?
I would think NMR would be too slow as a back-end detector for GC, and LC would be problematic because you'd need a large volume of deuterated eluent.  (Plus all the engineering problems with the need for a rapidly spinning sample.)
Title: Re: GC-MS versus NMR
Post by: DrCMS on October 21, 2013, 03:06:37 PM
@ Archer if you'd read his messages clearly and did not overlay your own expectation you might have realised curiouscat is not trying to ID a main product with some impurities.  He has already said the compounds he is trying to ID are in the 0.1%-5% range.  It is those impurities your are trying to tell him how to ignore that he is actually trying to ID so NMR of the crude reaction mix will not be very useful.
Title: Re: GC-MS versus NMR
Post by: Archer on October 21, 2013, 04:15:49 PM
@ Archer if you'd read his messages clearly and did not overlay your own expectation you might have realised curiouscat is not trying to ID a main product with some impurities.  He has already said the compounds he is trying to ID are in the 0.1%-5% range.  It is those impurities your are trying to tell him how to ignore that he is actually trying to ID so NMR of the crude reaction mix will not be very useful.

My apologies, I misunderstood.

DrCMS is absolutely right, the minor impurities need to each be purified to be the major components if NMR is going to provide any useful data.
Title: Re: GC-MS versus NMR
Post by: DrCMS on October 21, 2013, 05:31:26 PM
@Archer The thing to remember is that curiouscat is a chemical engineer involved with process improvement & control so identifying small impurities are important to work out how & why the yields are less than he wants.  Then he can work out if improving the yield is cost effective for that particular process.   

Your background is I think in analysis for regulatory control and proving a particular component is a controlled substance is your focus and so other minor impurities are not that important.

Those two roles are not really complementary.

My background is as a practical lab/plant chemist involved with process development & scale up so I can see where curiouscat is coming from which is why I posted the advice I did.  Then you posted the exact opposite which wound me up so I apologise if my reply to you seemed a bit sharp.  I can see from your reply to me that your posts were due to a minor comprehension error rather than a case of Maslow's hammer.
Title: Re: GC-MS versus NMR
Post by: curiouscat on October 21, 2013, 11:52:55 PM
@Archer @DrCMS:

Perhaps I added to the confusion because of notation: These are impurities with respect to my process, but OTOH for this analysis since they are the molecules of interest it may be a misnomer to call them an "impurity" except in the sense that they mess up each other's analysis by interference.
Title: Re: GC-MS versus NMR
Post by: Archer on October 22, 2013, 02:18:56 AM
@Archer The thing to remember is that curiouscat is a chemical engineer involved with process improvement & control so identifying small impurities are important to work out how & why the yields are less than he wants.  Then he can work out if improving the yield is cost effective for that particular process.   

Your background is I think in analysis for regulatory control and proving a particular component is a controlled substance is your focus and so other minor impurities are not that important.

Those two roles are not really complementary.

My background is as a practical lab/plant chemist involved with process development & scale up so I can see where curiouscat is coming from which is why I posted the advice I did.  Then you posted the exact opposite which wound me up so I apologise if my reply to you seemed a bit sharp.  I can see from your reply to me that your posts were due to a minor comprehension error rather than a case of Maslow's hammer.

No harm done, I was contradicting sound advice so your response was considered and polite relative to what you were probably thinking.

You and I have similar backgrounds in processing, my current role is the analysis you describe.

If we made something then minor impurities were things which were removed after the initial identification of the target molecule, then they would be identified at a later date after purification often used as standards for future process monitoring. One time an impurity turned out to be a valuable commodity.




Title: Re: GC-MS versus NMR
Post by: curiouscat on October 22, 2013, 02:29:01 AM
Quote
curiouscat is a chemical engineer involved with process improvement & control so identifying small impurities are important to work out how & why the yields are less than he wants.  Then he can work out if improving the yield is cost effective for that particular process.   

Couldn't have put it better. :)

In particular here's what happens. A process gets commercialized in a hurry and no one cares about what exactly is the 10%  that they form as by products.

Over time margins get squeezed and people start digging deeper. Unfortunately most often these side peaks have no characterization which makes it hard to try and reduce them in any systematic way.

Of course, often even after we know what they are there's no way to reduce them but that's another story. :)
Title: Re: GC-MS versus NMR
Post by: discodermolide on October 22, 2013, 04:17:28 AM
@Archer,
I was in the Pharma business and the impurities were and by the time I left still a valuable commodity. That was the state of affairs not very long ago.
Title: Re: GC-MS versus NMR
Post by: DrCMS on October 22, 2013, 04:35:39 AM
I know I can be rather harsh to people so I'll try to tone it down.  Thankfully as grown ups you and I hope me can take a bit of criticism without having a hissy fit and flaming everyone before running off to our mums. 

For what curiouscat is trying to do GC-MS is probably the best starting technique. 

If you could separate the components and get enough of a pure component NMR might give more structural info but the time and effort to do that purification and get an NMR run If you do not have in onsite might not be justified.  The GC-MS may well give you enough of an idea of the identity of the components of interest to figure out what side reactions are occurring and how you might reduce their formation.  You may well find that the cost of increasing the yield is more than the value of the extra yield.  A process we are fine tuning here just got the yield deliberately reduced because the cost in plant time and waste disposal of pushing the yield from 90% to 95% was costing more than the extra 5% of product sold for.  The people looking at process improvements had focused too much on the % yield rather than looking at the profit.  Once I did that is became quite clear that accepting the lower yield was a valid way to maximise the profits.