Hmmm... technique questions, kinda all over the place, but still, we're glad to help people learn. So, OK, point by point:
Hi there, although I can make a nice series of absorption spectra to determine the molar absorptivity, I would like to ask you for help for the following tecqniues problems (I would like to pursue most advanced tecqniues):
1. Is it desirable to warm up the machine (spectrophotometer) for 20 minutes before taking the measurement??
Warmup can be important, especially for older systems. But maybe not as critical. If you really care, do prepare test solution, and check it at various time points. Try to see how the instrument behaves. Then you can file a report, and have it on hand. So you'll have the answer for someone auditing your process.
2. Has anyone used the pipette ball? I know how to use it but I found the liquid inside keeps slowly sucking towards the bulb even though I stop pressing the "sucking region", which is very annoying...
Some people are experts at using those, but I'm not. At any rate, there should be a valve that stops the draw when released. I mean for the black bulb in this image: The red one would still draw, you break the seal in that case. Kinda wondering about the problems you're having, 'tho. Maybe you just need new bulbs, they do wear out over the years.
3. I found the baseline of my machine will shift down a very little bit (from -0.007 to -0.015) as I continue using the machine. Any explanation??
First off, is that significant for your application? Instruments do drift, it just happens. You'll have to explain if you're working at that low a level.
4. My solvent for measurement is DCM, you know, which is very volatile. Do you have any suggestion about how to minimize the error caused due to solvent evaporation??
Again, what how much error are you looking to avoid. Consider: you've prepared a 0.10 N solution and you get a reading of 0.5000 AU. The solution evaporates to a concentration of, oh I'll be crazy and say, 0.11. What absorbance would you get? You can calculate that using Beer's law. How much volume would you have to lose to get such a concentration, anyway? Measure the volume of a cuvette, I can only guess the path length, you have to figure out height of solution. If you're using DCM, that may be why you're having pipette problems. You may get more accuracy making solutions by weight.