HPLC problem (pressure fluctuate )
« on: April 12, 2006, 01:49:34 PM »
How can we solve the problem about the pressure in HPLC pump are not stable or fluctuate around 500 psi ?

  I used HPLC-UV (Water) to analyte. The problem will occur when I run the sample around 8-10 hour.

Offline kevins

  • Chemist
  • Full Member
  • *
  • Posts: 176
  • Mole Snacks: +17/-6
  • I'm a llama!
Re: HPLC problem (pressure fluctuate )
« Reply #1 on: May 16, 2006, 11:07:19 PM »
Please try to degas or purge the eluent.

Offline Moroni

  • Very New Member
  • *
  • Posts: 2
  • Mole Snacks: +0/-0
  • Gender: Male
Re: HPLC problem (pressure fluctuate )
« Reply #2 on: May 19, 2006, 07:33:51 AM »
Try to clean the bottle´s filters (from possibles fungis) and valves (from incrustantions of salts) with hot water (80ºC) + isopropilic alcohol  (80/20).

Moroni

Offline chiralic

  • Full Member
  • ****
  • Posts: 234
  • Mole Snacks: +19/-3
  • Gender: Male
  • Test
Re: HPLC problem (pressure fluctuate )
« Reply #3 on: May 31, 2006, 06:24:23 PM »
The causes of pressure changes in your HPLC equipment could be from mechanics and/or chemical origin.

If the Pressure changes, check the flow rate indirectly with measurement Dead Time (tm), in this way you can pinpoint the cause: FLOW or CHEMISTRY. Also you must check LEAKS.

If you have permanent small fluctations you must check:
+ Air bubles (in the pump, in pump head seal defect , frit of suction is clogged)
+ Pump head seal defect

If your pressure increases, deacreses, increases you must check:
+ Inlet valve is clogged (you must check pump head malfunction)

NOTE: Check in your system defect on pressure sensor

Regards,

Chiralic