Analytical procedure: Method detection limit and precision
« on: September 08, 2015, 11:53:37 PM »
Hello everyone.

Please i need clarification in simple terms on how method detection limit is obtained.

I have checked literature and quite contrasting or not clear statements to me were made.

For instance, the most popular one I see is "detection limit defined as 3σ (i.e 3 times standard deviation) of the measured lowest standard concentration (eg 10 nM) was ."   In order cases, they make mention of error bar.

I will appreciate if you can state in simple terms how to obtain the method detection limit of a particular analyte ,A, assuming I am using an HPLC-Fluorescence detector for quantification.

Precision:

I think I understand precision better. It is basically repeating a particular analysis using the same concentrations of reagents and reaction conditions several times and looking out for the variations in their response e.g. fluorescence signal intensity.
I will appreciate further insight on this also.

Thank you all
 

Offline Arkcon

  • Retired Staff
  • Sr. Member
  • *
  • Posts: 7367
  • Mole Snacks: +533/-147
Re: Analytical procedure: Method detection limit and precision
« Reply #1 on: September 09, 2015, 06:46:48 AM »
Hello everyone.

Please i need clarification in simple terms on how method detection limit is obtained.

I have checked literature and quite contrasting or not clear statements to me were made.

I can see that, because:

Quote
For instance, the most popular one I see is "detection limit defined as 3σ (i.e 3 times standard deviation) of the measured lowest standard concentration (eg 10 nM) was ."   In order cases, they make mention of error bar.

I assume that's a valid one that can be used, I've never seen it used before.

Quote
I will appreciate if you can state in simple terms how to obtain the method detection limit of a particular analyte ,A, assuming I am using an HPLC-Fluorescence detector for quantification.

Oh, that's easier.  For HPLC, the limit of detection is (often) defined as 3x noise.  You'll look at the signal height of the blank, or a blank area of the run, and use that as noise.  HPLC acquisition software will even do it for you in many cases

In fact, the method you've described above, 3 sigma the lowest slope, is used to define the Limit of Quantitation.
Quote

Precision:

I think I understand precision better. It is basically repeating a particular analysis using the same concentrations of reagents and reaction conditions several times and looking out for the variations in their response e.g. fluorescence signal intensity.
I will appreciate further insight on this also.

You can report that, but you may want to call that robustness, or something else, depending on the standards to which your analysis is held.
Hey, I'm not judging.  I just like to shoot straight.  I'm a man of science.

Offline JGK

  • Chemist
  • Full Member
  • *
  • Posts: 738
  • Mole Snacks: +66/-19
  • Gender: Male
Re: Analytical procedure: Method detection limit and precision
« Reply #2 on: September 10, 2015, 06:18:30 PM »
I would suggest you look at the ICH guideline on validation sections 6 and 5 respectively.

http://www.ich.org/fileadmin/Public_Web_Site/ICH_Products/Guidelines/Quality/Q2_R1/Step4/Q2_R1__Guideline.pdf
Experience is something you don't get until just after you need it.

Offline marquis

  • Full Member
  • ****
  • Posts: 393
  • Mole Snacks: +35/-3
Re: Analytical procedure: Method detection limit and precision
« Reply #3 on: September 12, 2015, 02:29:37 PM »
Dumb question-

Last time, there was still two paths for American industry, either tripartite of ISO.

Has the mess settled?  Is one standard now?

Thanks,

Sponsored Links